here’s an interesting graph I’d like some feedback on.

dark purple: Right VL
light purple: left VL

Concept 2 BikeErg 60min continuous ride.

Around minute 24 RVL Moxy dies (battery) so I switch it rapidly and use another one.

i was wearing the prototype moxy shorts so the placement was VERY close to where the previous moxy was (at the very least).

surprisingly the absolute % is quite different from one to the other moxy! Although the trend is very similar.

now which is "right"? The first or the second moxy? Depending on this answer it would change the interpretation of the relative performance between Left and Right.

what do you think is going on here? Wondering what the implications are for interpreting L/R differences and whether or not that can be done (on absolute values that is, the relative difference in slope seem very close here).

what is your take on this?
thanks in advance!
andrifeldmann
Hi Sean, we (I) have done a lot of experimenting with changing sensors, and find that the signals are very consistent. Changing sensors during activity is quite hard, as you do have to stop, remove the sensor and replace the sensor. I assume you had to do this. The stop will result in recovery of smo2, so it will shift up. Then when you start riding again, you start the ride at the same intensity, but with a local recovery, at the same time your cardiopulmonary system is still ramped up (assuming the break wasn't that long), which means that your overall smo2 levels may be higher due to the recovery, followed by same vo2 after. Does this make sense? Interesting is that the break to swap the moxys didnt results in any change in the other leg? Your concern about absolute R/L evaluation is of course warranted. Try to duplicate if this happens again. 
SeanUpsideStrength
Thanks for your thoughts Andri! I actually kept riding since I had the shorts on - notices the loss of signal and just kept going, popped it out and out the other one in. So no stop that could explain the hike in SmO2
bobbyjobling
Contact pressure has an effect on the reading. I saw an article about this a few years ago where they compared a moxy to another device.
MoxyPhysiology
Greetings Sean, 

Something to note: The graphs have different scales (e.g. Moxy 1 ranges from 20-60%, while Moxy 2 ranges from 40-60) this could make the difference seem a lot higher than it truly is. Just from a rough eye test, I am guessing that the difference in SmO2 is around 3-5%. I haven't done as much testing as Andri, but I think that 3-5% could be within the realm of variability between devices? 
SeanUpsideStrength
Good eye! I will pull the CSV out and graph it to see what comes out. Thanks!
andrifeldmann
I recommend always graphing smo2 on a 0-100% scale, and not adjusting this. Scale mistakes are easy. And because the 0-100% scale is relevant for SmO2, it should always remain. 
bobbyjobling
If you still have a difference after the scale is adjusted you might want to read this.

Testing the Influence of External Pressure.
There were significant effects of the applied pressure on restingTSI, especially for the MOXY. At low external pressures (2 to20 mmHg), the PortaMon reported a lower resting TSI (∼8%)compared with the MOXY. As external pressure was increasedabove 20 mmHg, the difference between the two devices dimin-ished. This difference was predominantly due to changes in theMOXY value, not the PortaMon. In particular, MOXY valuesshowed a sharp decrease between 20 and 30 mm Hg. As externalpressure is applied to the skins surface, arteriole, venule, andcapillary intravascular hydrostatic pressure are equal in relationto the distance from the surface of the skin,34 occluding the flow of blood to the periphery. As a larger proportion of light detectedin the MOXY comes from surface tissue it is likely that theMOXY is more sensitive to pressure-induced changes in the der-mis and subcutaneous oxygenation and blood flow. Again, theSR algorithm used to measure muscle TSI is less effective thanthe SRS method in the PortaMon in correcting for the effects ofdifferences in blood flow and oxygenation in surface tissues.This finding has an important implication for researchers whenundertaking repeated measures using the MOXY device,particularly when the removal and reapplication of the NIRSdevice is required. Furthermore, athletes using similar devicesto inform training intensity should standardize the external pressure applied to hold the instrument in place

https://www.researchgate.net/publication/322690563_Performance_comparison_of_the_MOXY_and_PortaMon_near-infrared_spectroscopy_muscle_oximeters_at_rest_and_during_exercise


andrifeldmann
I have talked to Prof. Cooper about this part of there paper, as I could not replicate this response to the same degree. Have you tried replicating this? I get very minimal difference in Smo2 with significant changes in external pressure. I never actually measured the amount of change in applied pressure, so this is subjective, but I made a video about this:

https://www.youtube.com/watch?v=uOKov3JBNqo

Maybe Roger can comment from a technical perspective. 
andrifeldmann
Sean, can you give us more of a play-by-play of exactly what you did, it looks like a 5% swing with the sensor change. This could be accumulating error. Maybe pressure as mentioned by bobbyjobling, and slight position change, and some physiological shift, as there must have been some slow down in pedaling and drop in power. What do you think? Another idea is to try to repeat the happening. 
vo2max
I've noticed changes in Smo2 and THb with minute shifts of the sensor position on the leg.  So I would not count out a slightly different position as contributing to some of the difference.
SeanUpsideStrength
Here's the same data with the CSV exported on a single graph ( instead of the two separate scales on the VO2 Master report).

What do you think? What should I do next? HR[bpm], SmO2[%], SmO2 - 2[%], SmO2 - 3[%], THb[THb]….png
SeanUpsideStrength
@andrifeldmann to answer your question, I saw that the graph went to zero on the VO2 Master Manager App, popped the Moxy off the shorts (it was off - battery dead), pulled another Moxy from my box and put it in place of the first one. Then turned it on and synced it to the Moxy data stream #1 again so that it would show up on my app. I don't remember taking my leg off the bike but as you said maybe I was going easy on that side while I did the change. But the "recovery" between the Moxy with an empty battery and the replacement is pretty substantial still!

Happy to conduct a follow up test - let me know what to do! I can try switching around multiple Moxys if needed.
andrifeldmann
if you have time maybe see if you can reproduce this. It is an unexpected response. @vo2max, of course small shifts can result in smo2 change, but they are generally very small, to negligible. But it could be an accumulating interference event; position/pressure/physiology.
andrifeldmann
One more point, in regard to right and left balance. You should never make a definite decision of a right and left balance based on a single data set. The difference needs to be consistent over time and exercise. 