His SmO2 values are very low, remember that NIRS has some limitation in quantification (something that is in debate, the quantifiable is usually the discussion between NIRS devices, whereas trending is generally the same). So if one athlete has an SMO2 of 60 and another 70, it may be a very similar value, however with very large differences and very low numbers such as what you see here it is more likely than not that he does actually have a low muscle saturation. Just to consider.
Then secondly, considering these low values, it would seem that there is an imbalance between o2 supply and utilization and this can be a result of a variety of combinations.
Poor supply and good utilization
Good supply, and really good utilization
Very poor supply, and poor utilization
And so on, as you can see it depends on the relationship between utilization and supply.
My add on for the moment here.
A step test has some problems to actually assess the above open questions and possibilities. If we have a steady increase in load ( muscle compression ) as well a steady increase n potential delivery of O2 as well as utilization of O2, than it is very difficult to see what is going on.
If we suddenly interrupt a load ( 1 min rest ) we stop immediately the O2 demand or better energy demand ) but we do not stop immediately supply line Cardiac and respiration still up )
so we in simple terms see how " much " O2 was used by muscle activity. ( overshoot or spike in SmO2 at the rest.
On the other side of the 1 min when we have dropped Cardiac and respiratory a activity and as such lower delivery but a sudden high demand so the downwards spike is an indication of immediate utilization due to need of energy but not yet optimal delivery.
Same holds true for blood flow indication.
mechanical reasoning of dropping or increasing tHb is fast so again in the break we see the reaction of tHb and have an indication on how much muscle mechanical compression may have reduced blood flow tHb or may reduce blood flow tHb due to new start.. Keep this in mind when we go back to this case.
So when you look at this reactions look at the levels of SmO2 in the rest period over the full assessment time as well the level of tHb in the rest period.
See in the different pics up and you see, what the pic indicate after what we tell you here now.
In order to determine how good the utilization is you need to determine how good the supply is, and one very simple way of doing this is to also measures spo2, which gives an indication of arterial blood o2 saturation. If this is already below average there is some kind of supply problem, likely from a respiration deficiency.
Just some thoughts. ( Andri )
Last but not least. An increase in tHb but a flat SmO2 or even drop in SmO2 at the rest period can indicate some very interesting situations. See here agin a so called expected "normal " reaction when looking at HHB and O2Hb ( HHB is blue and O2Hb is red. and a " unusual HHb O2Hb reaction.
remember tHb is O2Hb + HHB and SmO2 is the % of O2Hb to tHb
Here the "unusual" picture
And here just to answer some expected mails for MOXY questions.
Yes MOXY shows the same if the situationis the same.
Some critics argued, that in MOXY if O2Hb drops HHB will go up and vica versa and we never would see an increase in HHb and O2Hb. We sent thsi critics many cases and never got a retunr fo discussion. In short critic is great whne we can discuss teh issue.
Here a MOXY assessment with the above " unusual " sitution but once you understand the basic physiology you understand why this can happen.
look at SmO2 and tHb at the end.
and just to avoid a flood of emails.
Yes we can have as above an increase in tHb but a drop in SmO2 as there is a very different situation why that happens. here an assessment example
Immagine the SmO2 trend of the trace ? yes here it is.