First, it is important to understand some rules and limitations of NIRS and Moxy in regard to the data you are collecting. When we are comparing a measured value, like your difference in RF and VL we need to be careful as the margin of error is larger here and we could make an alpha error. Within 10% differences we do not consider a difference. So in the case of RF and VL, or right and left, if the numbers are within 10% (51% and 60% for example), this is looked at as the same. Once the range is bigger than 10% now we have increasing confidence that there is actually some physiology to differentiate, and would start to look at some different data to see if we get some additional information. For example in a left right comparison, are the left and right both producing the same power? That being said, about the value, comparing trends are much more robust. Which means if you have one measurement down trending by 5%, and another by 9%, this is much more accurate to compare, as a great down trend in B than in A. What you can see in your example, is that generally all three test have very similar trends (two are the same so they should). If you take a look at the Andrew Coggan webinar he shows some data that identifies same trending across a wide range of lower body muscles when cycling. I hope that is a first step.
This goes directly into the question of SmO2% and low values, can that be? Well its hard to get a solid answer as muscle capillary measurements in vivo and in situ is rather difficult. However, generally values under 20% are excepted, especially looking at SmO2 rather than StO2. If a 0% values is really 0%? Likely not, gain we consider the 10% range. But a 0%, or 3% or 5%, just means very very low. It means you have essentially reached a maximally deoxygenated state where you are extracted and utilizing all the supplied O2. In many ways with NIRS we talk much more about maximally deoxygenated states, rather than an absolute % of SmO2 as this is difficult to prove.
For tHb, we do not compare the values at all, only the trending.
Having hopefully clarified these points, I look forward to your questions, closer to type 4 you mention. Ill give a quick insight that may help. If you look at tHb in your 3rd RF test and compare that to your 1st and 2nd VL test you can see that the tHb trending is a little different. You can see that in the RF test tHb is a little more dynamic and moves a little with SMO2 and at the end increases. tHb is an indirect measure of blood volume changes. This means that you have a more dynamic blood volume change in the RF than in the VL. This of course could then account for higher SmO2 values and slightly different trends.